TEMPERATURE RELATED QUESTIONS
- Are the GB4 and GB5 Sending Units different?
Answer: No; they're the same. And so is the optional GB1 Sending Unit.
However, Sending Units differ by thermocouple type. All Sending
Units for Type K Thermocouples are the same, but they are different
from the Sending Units for Platinum Thermocouples (Type R or Type S).
- Why do Sending Units have shipping jumpers?
Answer: To protect against possible damage from static electricity during
shipping.
- All of a sudden one of my ovens reads 32ºF. Why?
Answer: Assuming the other channels are working correctly, the most likely
cause is that one or more of the wires between the Sending Unit and
the GB4 or GB5 is either broken or disconnected. A less likely
possibility is that the Sending Unit itself is defective.
- All of my ovens read very low, about 34ºF to 45ºF. Why?
Answer: A GB4 or GB5 has an internal power supply for its Sending Units.
This power supply protects itself from external short circuits by
shutting down almost completely, causing a low reading on all
channels, even though only one may be shorted out. Since this
reading is below any reasonable setpoint, a running unit would remain
on continuously, trying to attain its setpoint. Left unchecked, this
could cause a serious over temperature situation:
A GB4 prevents this potentially dangerous condition by refusing
to allow a unit to run if the temperature is below about 50ºF.
A GB5 handles this in a more sophisticated manner: it reports a "BAD1" error and
shuts off the corresponding unit.
- What could cause the short circuit you mentioned above?
Answer: The most likely cause is a defective Sending Unit. Another
possibility is that the insulation of the wires is damaged, allowing
bare metal to touch something it should not.
- My GB4 controller was working fine all summer and fall. Last
night it got quite cold in the barn where my annealers are. In
the morning none of my annealers would come on. Why?
Answer: The GB4 interprets the cold temperature in your barn as the
potentially dangerous problem mentioned above. Raise the
temperature reading slightly to "jump start" you unit by holding the
thermocouple in your hand or heating it briefly with a flame. Once
it's above 50ºF, it will work normally. Do this for each channel that won't come on.
- Why do my temperature readings jump around?
Answer: Likely possibilities:
- bad theromcouple, or
- a loose connection between the thermocouple and the Sending
Unit, or
- the "multiplexor" inside the GB4 or GB5 needs replacement.
If the problem occurs on more than one channel, it is almost certainly
the multiplexor.
- What's a multiplexor?
Answer: This is an integrated circuit that connects each sending unit in
turn to the computer circuitry. Since it is electrically close to the
input, it is more sensitive electrical damage than most of the
other parts of the controller. These include static electricty,
touching a thermocouple to heating elements, lightning strikes in
the neighborhood, etc.
- What is a pyrometer?
Answer: It's an instrument used for measuring high temperatures.
Sometimes "pyrometer" is used colloquially to refer to just a
thermocouple probe instead of an entire instrument.
- What is a thermocouple?
Answer: A temperature probe based on the contact of two dissimilar metals.
Any time two dissimilar metals are in contact, they generate an
extremely small voltage that depends on the temperature of their
junction. It also depends on what the metals are. To be useful, the
metals must have certain characteristics and the correspondence
between the temperature and the voltage it generates must be known.
- How is a thermocouple different from a thermocoupler?
Answer: A thermocoupler is simply a mis-pronounced thermocouple.
- What does "Type K" mean?
Answer: Type K thermocouples are one of the common, standard
thermocouples. They use Chromel (a special alloy of Chromium and
Nickel) and Alumel (a special alloy of Aluminium and Nickel) as the
two dissimilar metals to generate the voltage. This pair of metals
has been studied extensively and the temperature-voltage
correspondence is well known.
- Are there any other types of standard thermocouples?
Answer: Yes, there are several, each based on a different pair of metals, and
designated by a letter of the alphabet. For example, Type J uses
iron and an alloy of copper and nickel; Types R and S use slightly
different alloys of platinum and rhodium as one of the metals and
platinum as the other. Different metals impart different properties
to the various types of thermocouples, including the maximum
temperature they can withstand, the strength of the voltage output,
etc. For example, Type J does not hold up well at temperatures above
1300ºF.
- Why do you use Type K thermocouples? What other ones could I use?
Answer: For working with glass, Type K thermocouples are the de facto
standard. They are relatively inexpensive, physically rugged,
and easily handle most of the temperatures involved with fusing,
slumping, kiln-casting, annealing, etc.
You could also use Type R or Type S thermocouples, which withstand
even higher temperatures than Type K, but being made of platinum,
they are much more expensive and are generally reserved for high
temperature applications such as furnaces. Also, to keep their cost
down, platinum thermocouples are made of very thin wire and thus are
very fragile. At any given temperature, they produce a smaller
voltage, requiring more costly electronics.
- Can I use any type of thermocouple with a Digitry controller?
Answer: Digitry controllers are calibrated for only Type K, Type R, or Type S
thermocouples. You must specify the type of thermocouple to be
used for each channel when you order the controller.
- What is the difference between type R and Type S thermocouples?
Answer: There is very little difference between the two. We don't know why
both exist. They both use the same kind of Sending Unit, but they do
require slightly different calibration, so when ordering a
controller, you must be clear which one you will be using. Type S
seems to be more common than Type R.
- Why shouldn't I always use Type S themocouples? It seems they
might last longer?
Answer: As noted above,
Type S thermocouples are much more expensive
and fragile than Type K thermocouples. You can buy many Type K
thermocouples for the cost of a Type S thermocouple.
- Do I need special wire for connecting my thermocouple?
Answer: You should use special thermocouple extension wire to connect a
thermocouple to a Sending Unit or to a GB1. You do not need
special extension wire to connect the Sending Unit to the GB4 or
GB5. A popular Type K thermocouple sold by Digitry includes an
integral 7' lead wire, so no extra extension wire is required.
- What is the difference is between your power surge supressor and one
from [name your favorite store]? Is there a drastic difference in quality?
Answer: YES! Our equipment is based on silicon avalanche suppressor diodes (SASDs)
rather than the older technology of metal oxide varistors (MOVs) that most
units use. SASDs have far better performance for modern electronic devices and
a much longer expected lifetime, something like 10+ years. An MOV degrades
over time and can be used up in a matter of days or years, depending on the
number and intensity of transient voltages ("spikes") it
experiences.
- As a first quick test, to be sure everything was hooked up correctly,
we did an initial run up to 500ºF degrees in 5 minutes. Why did it overshot to 576ºF?
Answer: The GB calculates the required temperature for each minute, as determined
by the profile you enter. If the observed temperature is below that, the
standard GB calls for heat by turning the contactor on; if above, it turns
the contactor off. When ramping up, when the kiln reaches the desired
temperature and the contactor turns off, the residual heat in the heating
elements will cause it to continue to heat. How much it heats depends on a
lot of things, like how big the kiln is, how powerful the elements are, how
well insulated the kiln is, what's in it, and so on.
Whenever the kiln is going full blast for some while, the elements
get to their hottest temperature, maximizing the overshoot. When the
kiln is near the set point and the elements come on to merely correct
for heat loss or to ramp up to a nearby value, they usually do not
reach their hottest potential before the required temperature is
attained. This reduces the overshoot. The situation you described
will probably have the kiln on constantly until it gets to 500ºF
degrees. Hence you have a lot of residual heat and more overshoot.
If your program has a second step, one that holds the kiln at 500ºF
degrees, the subsequent temperature swings will be much less. If
there were a large piece in the kiln (thus increasing the "thermal
mass"), it would be even less. We do have proportioning models that
reduce the power to the elements when the temperature gets close to the
target, thus reducing the overshoot.
|